PostPosted: Sat Sep 22, 2007 8:40 pm    Post subject: Retrospective reconstruction in CT? Reply with quote

I've done a fair amount of reading on this but i'm still not sure.
So if any of you are able to help me with this, it would be fantastic!

I know that in multi-slice CT, once we have performed a scan, we are able to retrospectively reconstruct images with different slice thickness from the existing raw data without re-scanning the patient. But if you want to reconstruct a THINNER SLICE...the thickness of this slice is limited by the detector configuration that was selected.

From what I understand, you cannot retro-reconstruct smaller than the smallest detector width which was employed.
Is that right?
Could someone please clarify this for me.

If the detector configuration was 2x2.5mm can you reconstruct a 1.5mm slice??

any responses will be much appreciated. cheers!

ScrmnWoody



Joined: 22 Aug 2007
Posts: 40
Location: Bellingham, WA

PostPosted: Sun Sep 23, 2007 11:29 am    Post subject: Reply with quote

I'm a little confused by your question. Here's why. We have a Toshiba Aquillion 32 slice scanner. Almost everything we scan (if helical) is scanned at 0.5x0.3mm.

Most things are reconstructed prospectively at 3mm. But, if we needed to go back and retrospectively reconstruct something smaller, we could. And, occasionally the radiologist will simply ask us to send the volume to PACS and they can review the 0.5x0.3.

It doesn't get much smaller than that.
_________________
Chris

rekex



Joined: 05 Mar 2006
Posts: 19

PostPosted: Mon Sep 24, 2007 3:55 am    Post subject: Reply with quote

@ ScrmnWoody
Thanks 4 ur response

but i suppose i should clarify my question. What i am asking is:
For example with the 4-channel Siemens helical scanner, the detector array is designed as follows:
5mm, 2.5mm, 1.5mm, 1mm, 1mm, 1,5mm, 2.5mm, 5mm

If the detector configuration of 4 x 5mm is used. So the 2.5mm, 1.5mm & 1mm elements are grouped to form 5mm width.

then in this case we are acquiring 4 slices of 5mm section width. and if we wished to retrospectively reconstruct thinner slices, say 2.5mm or 1.5mm would it be possible?

From what I have read it seems that we cannot reconstruct any smaller than the effective detector thickness. But if we wanted to reconstruct 10mm in this case it would be possible. Is this correct??

I hope I've explained my question clearly.

papa
Power User


Joined: 02 Oct 2007
Posts: 211
Location: The State of Confusion

PostPosted: Fri Oct 12, 2007 3:23 pm    Post subject: Reply with quote

sometimes it depends on the scanner. so you've got the seimens 4-slice, and you're trying to take 5mm slices and change them to 2.5mm slices. you need to scan the images in the 2.5mm slices. the scanner can fuse them together into a biger slice, but you can only reconstruct down to your smaller slice. so if you want 2.5mm slices, you need to change your set-up to 4 X 2.5 fused to 5.0mm or 2 X 5.0.

the reason you can't reconstruct those 2.5, 1.5, and 1.0 mm slices is that the entire scan needs to be performed in the smallest slice. the scanner can't take just those images and reconstruct smaller, because you would have image gaps all over the place.

another question is if you are scanning helically or axial. the scanner does both. if you are scanning helically, to get a thinner slice, make sure the "narrow collimation" is checked to on. then you can reconstruct down to like .75mm thick.
_________________
Mr. Green papa RT(R)(CT) Mr. Green

http://sites.google.com/site/delusionalradiology

never meddle in the affairs of a dragon, for you are crunchy, and taste good with ketchup

rekex



Joined: 05 Mar 2006
Posts: 19

PostPosted: Tue Oct 16, 2007 9:36 pm    Post subject: Reply with quote

@ papa

Thanks for the response. i think i understand now Smile

Post new topic   Reply to topic
Page 1 of 1
View previous topic :: View next topic

 
Jump to:  
You cannot post new topics in this forum
You cannot reply to topics in this forum
You cannot edit your posts in this forum
You cannot delete your posts in this forum
You cannot vote in polls in this forum



Powered by phpBB
Xray Forums Archive - Radiology Jobs - Radiology Schools - Continuing Education - Online Nursing Schools

ScriptWiz.com phpbb HTML Archiver - Created by ScriptWiz.com and released by Skinz.org