Radiology Discussion Boards

Google

Welcome Visitor, Please Register or Login
Post new topic   Reply to topic
Author
papa
Power User


Joined: 02 Oct 2007
Posts: 211
Location: The State of Confusion

PostPosted: Fri Nov 02, 2007 11:04 am    Post subject: physics testing on GE 64 Reply with quote

ok, dumb questions.....

we had the physics testing for the ACR accredidation on the GE 64. the physicist failed us on the low contrast resolution, and a ring artifact. he told me that since this is a brand new scanner, he would expect better image quality from it.

had my GE service engineer come out to recalibrate the scanner, and he told me that everything is within GE specs.
-the phantom that the physicists use for ACR won't give the propper images, and they need to use the image analysis tool for the low contrast resolution.
-the ring artifact is from scanning the phantom helically at a smaller DFOV. if he were to test it at a larger FOV, he would not see the ring.

so who do i believe, GE or the physicist?

Question
_________________
Mr. Green papa RT(R)(CT) Mr. Green

http://sites.google.com/site/delusionalradiology

never meddle in the affairs of a dragon, for you are crunchy, and taste good with ketchup

papa
Power User


Joined: 02 Oct 2007
Posts: 211
Location: The State of Confusion

PostPosted: Mon Nov 05, 2007 2:57 pm    Post subject: Reply with quote

ok, another thing wrong ( i always thought GE was better than this)

another failed part of the ACR was the HU for water. when i had the GE service engineer in, we tried to duplicate the problem, but could not tell what values the physicist was using (helical/axial etc.)

well, today I just scanned every damn thing. I found out that on my 100 kV 5.0 mm axial scans, i was getting like a -6.0 for water, and on my 120 kV 1.25 mm images, they were going as far out as -7.2.

told this to my service engineer, and he said that he would check with his supervisor, and that this may be normal for the "new technology in the GE VCT 64".

who thinks i'm getting a hose job?


Evil or Very Mad Exclamation
_________________
Mr. Green papa RT(R)(CT) Mr. Green

http://sites.google.com/site/delusionalradiology

never meddle in the affairs of a dragon, for you are crunchy, and taste good with ketchup

CigarPete
Power User


Joined: 21 May 2007
Posts: 89

PostPosted: Tue Nov 13, 2007 9:56 am    Post subject: hmmm Reply with quote

well papa..

welcome to the world of the middleman. I would get the physicist in touch with the service guy.. and let them hash it out... a lot of physicists are smarter than this tho..

I would never let the service guys yank me around.. I show them the report and say tell them this needs to be fixed.. GE is famous for tihs stuff... if there are issues getting the service guy to fix what's wrong then ask him for the name of his supervisor and call that guy.. If GE has one thing it's certainly personel in management roles.

I hate these kinds of fights..

Good Luck
Pete

Post new topic   Reply to topic
Page 1 of 1
View previous topic :: View next topic

 
Jump to:  
You cannot post new topics in this forum
You cannot reply to topics in this forum
You cannot edit your posts in this forum
You cannot delete your posts in this forum
You cannot vote in polls in this forum



Powered by phpBB
Xray Forums Archive - Radiology Jobs - Radiology Schools - Continuing Education - Online Nursing Schools

ScriptWiz.com phpbb HTML Archiver - Created by ScriptWiz.com and released by Skinz.org