PostPosted: Tue Oct 06, 2009 10:46 am    Post subject: Help understanding ionization chambers Reply with quote

Hello. I'm just looking for some help with regards to understanding ionization chambers. I've checked out google etc, but just wondered if someone could see if I'm on the right track here...

Scenario

You're about to perform an erect PA chest X-Ray. You select the kV and mAs, and take the X-Ray. Would I be right in saying that the ionization chamber(s) lie behind/are built into the bucky/behind the image receptor somewhere, so that once the ionization chamber(s) detect the kV and mAs that you set for the X-Ray, it then cuts off somehow so that the patient does not receive more kV or mAs than was originally set by yourself for the examination?

Just a little confused, and need this clearing up, if anyone can help? Very Happy

wvaio



Joined: 22 Sep 2007
Posts: 82

PostPosted: Tue Oct 06, 2009 7:20 pm    Post subject: Reply with quote

Ahh... I can explain this one for you!

Layman's version:

Directly behind the grid on the wall bucky there are three chambers. They basically count the number of photons that are going to hit the cassette. When a determined number of photons (enough to produce a certain density on the image) are produced they tell the machine to stop. If they don't count enough x-rays before the back-up timer goes off, the timer terminates the exposure. The chambers will count any type of photons including scatter. That's why it is necessary to collimate and place the part over the correct cells. You mainly want them to count the photons in the area of interest. If they read too many photons from unnecessary exposure, you can end up with quantum mottle. The chambers do not care what kVp you use. They only care about ionizations produced by photons.

Hope that helps.
Wvaio

traineetech



Joined: 06 Oct 2009
Posts: 15

PostPosted: Wed Oct 07, 2009 5:25 am    Post subject: Reply with quote

thank you! that's a great help!

traineetech



Joined: 06 Oct 2009
Posts: 15

PostPosted: Wed Oct 07, 2009 5:34 am    Post subject: Reply with quote

P.S So there'd also be 3 chambers on the bucky table when carrying out an X-ray of say, an abdomen when the patient is lying down?

So, just to clarify, an ionization chamber is bascially there to makre sure that no more mAs/photons reach the image receptor than originally set by the technician?

IE, if you set an mAs of 15, once the ion chambers detects 15mAs, the exposure is stopped?

traineetech



Joined: 06 Oct 2009
Posts: 15

PostPosted: Wed Oct 07, 2009 5:39 am    Post subject: Reply with quote

P.P.S another thing which has been bugging me...

When referring to a bucky, I thought that a bucky was the same thing as a grid as in Potter Bucky Grid? or am I getting this wrong?

Just got confused when you said "Directly behind the grid on the wall bucky there are three chambers"

wvaio



Joined: 22 Sep 2007
Posts: 82

PostPosted: Wed Oct 07, 2009 8:19 pm    Post subject: Step two Reply with quote

Now that you get the general idea, I will up the ante a bit. I had the same questions when I started clinicals. It took me a while to figure out exactly what was happening. Its good that you are asking these questions. Understanding the AEC system is important to performing well in this field.

Here it is:

When you set a manual technique,as in 55kVp at 3.2 mas, the xray machine will shoot 3.2 mAs at 55kilovolts PEAK. That is calculated by the mA x the time.
The ionization chambers come into play when you use AEC(Automatic Exposure Control). The AEC system is ran by the computer system of the unit. Its function is to create a predetermined density on the cassette placed into the Bucky tray. The AEC uses the ionization chambers to calculate that exposure. When more than one chamber is used, the results of all chambers are averaged.
Lets say you are shooting a KUB. The patient is supine on the table. At a lot of places the computer would do the following steps for you just by pressing Abdomen table bucky, but for the sake of this example you are setting the factors by hand. First, you would choose a kVp. Lets say 80 kVp. Then you would select the number of chambers and their position. You would want to use all three chambers because the area of interest will completely cover all of the chambers. Next you would pick the mA that you want to use. We'll choose 240mA. When you press the exposure button, the tube will start to produce X-rays at 80kVp at 240 mA. The ionization chambers will start counting the number of photons that pass through the patient and the ones scattered by the patient. When the computer thinks the cassette has had enough exposure, it will terminate the exposure. The computer has a back-up timer that will terminate the exposure if it takes too long for the chambers to read their required number of photons. This protects the patient from being over-exposed due to a system failure( or operator error). If the machine produced x-rays at 80 kVp at 240 mA, and the ionization chambers calculate that they have the necessary exposure after .167 seconds, the machine will have produced 40 mAs. You could then turn off the AEC and set a manual technique of 40 mAs at 80 kVp, and achieve the same image as you did with the AEC on that same patient.

Hope that helps,

Wvaio

traineetech



Joined: 06 Oct 2009
Posts: 15

PostPosted: Thu Oct 08, 2009 6:32 am    Post subject: Reply with quote

Hi again. Yes, I think that has cleared it up for me, so thanks for writing that! I'll go over it again in my head later, and if any further questions, shall gvet back to you.

Thanks!

Post new topic   Reply to topic
Page 1 of 1
View previous topic :: View next topic

 
Jump to:  
You cannot post new topics in this forum
You cannot reply to topics in this forum
You cannot edit your posts in this forum
You cannot delete your posts in this forum
You cannot vote in polls in this forum



Powered by phpBB
Xray Forums Archive - Radiology Jobs - Radiology Schools - Continuing Education - Online Nursing Schools Alabama Radiology Schools - Alaska Radiology Schools - Arizona Radiology Schools - Arkansas Radiology Schools - California Radiology Schools - Colorado Radiology Schools - Columbia (D.C.) Radiology Schools - Connecticut Radiology Schools - Delaware Radiology Schools - Florida Radiology Schools - Georgia Radiology Schools - Hawaii Radiology Schools - Idaho Radiology Schools - Illinois Radiology Schools - Indiana Radiology Schools - Iowa Radiology Schools - Kansas Radiology Schools - Kentucky Radiology Schools - Louisiana Radiology Schools - Maine Radiology Schools - Maryland Radiology Schools - Massachusetts Radiology Schools - Michigan Radiology Schools - Minnesota Radiology Schools - Mississippi Radiology Schools - Missouri Radiology Schools - Montana Radiology Schools - Nebraska Radiology Schools - Nevada Radiology Schools - New Hampshire Radiology Schools - New Jersey Radiology Schools - New Mexico Radiology Schools - New York Radiology Schools - North Carolina Radiology Schools - North Dakota Radiology Schools - Ohio Radiology Schools - Oklahoma Radiology Schools - Oregon Radiology Schools - Puerto Rico Radiology Schools - Pennsylvania Radiology Schools - Rhode Island Radiology Schools - South Carolina Radiology Schools - South Dakota Radiology Schools - Tennessee Radiology Schools - Texas Radiology Schools - Utah Radiology Schools - Vermont Radiology Schools - Virginia Radiology Schools - Washington Radiology Schools - West Virginia Radiology Schools - Wisconsin Radiology Schools - Wyoming Radiology Schools

ScriptWiz.com phpbb HTML Archiver - Created by ScriptWiz.com and released by Skinz.org